G'day Leigh,
This is a not uncomon problem with the the GSO solid tube dobs. I have carried out this maintenance task on 4 of them over the last couple of years. My "20 cents worth" for your consideration

:
1. Yes, regardless of how you tackle this task, you should get a collimation tool and become familiar with Newtonian collimation. This will help you get the best performance out of your scope. With a bit of practice, collimation is easy and fast.
2. I suggest the safest way to approach this task is to remove the primary mirror.
3. To avoid having to repeat this task, you should consider applying some "Loctite" to the bolt threads prior to retightening. if you decide to do this, you will need to completely undo all 4 altitude trunion bolts.
4. To remove the mirror (in its cell), you need to:
- Prepare a padded flat surface of sufficient size for the telescope tube and the mirror cell laid flat behind it.
- Have some rolled up towels or padding to place along side the tube to stop it rolling.
- Have somewhere safe to place the mirror and cell while you are working on the tube. This should have a cover of some sort to avoid accidents - a very large cake tin or carboard box etc. This shold be placed well away from the tube while you are working on it.
- Have a small container for the screws that hold in the mirror cell.
- If you haven't done this sort of job before, have at least 90 min free - don't rush.
- Lay the tube horizontal on the padded surface and place the extra padding along both sides to stop it rolling.
- Using a couple of pieces of stick tape, or similar, mark the tube and rear of mirror cell so you can replace it in the same orientation.
- Remove the 6 primary mirror cell attachment screws and put them in the small container.
- Have an assistant hold the telescope tube and carefully pull the mirror cell out. This may require a bit of force as it may be quite a tight fit - make sure you are not working at the edge of the table for this in case the cell comes out in a "hurry"
.
- Place the mirror and cell in a safe place and you should now be able to easily reach the trunion attachment bolts for maintenance as required.
- Once you have tightened the bolts up, reverse the above procedure, collimate the scope and you are back in business.
Regards,