Marc if the second camera is not connected then there should be no reading in that second panel at all. I am a little confused as to why it is there. Yes by the description you gave of the TEC behaviour something is definately amiss. It takes mine about 5 mins to equalise at -20 and generally moves between -19.9 to -20.1 at the absolute most but only as the TEC % backs off during the night as it gets colder and it is hunting for the new setting. I am sure Quiy and Theo know what they are doing and will have you sorted quickly  .
Mark
|
I have no idea why either but not worried about it at this stage. One thing at a time. Theo should be sending me another DC201 soon to test again. 1% on the TEC power slider will bring the temp to -30c so it's all or nothing. It doesn't regulate. 1% or 70% has the same effect and I think that's where the problem lies.