Quote:
Originally Posted by Merlin66 View Post
Ahhhh
You may have found the problem/ issue!
Those two screws you show in photo#4 should NOT be there!!
It looks like the screws, used to attach a finder etc are toooooo long and probably interfere with the mirror travel.
They should be unscrewed to at least stop any interference with the mirror.
Quote:
Originally Posted by casstony View Post
Was the tube originally in a fork? Looks like a deforked model and the screws used to fill the holes are too long.

Thanks. Those screws are definitely in the position where the forks attach to the tube. They have always been there since I bought the scope new in 2010!! The focuser had always been easier to turn in some places and harder in others. I always thought it was just a focuser thing. The scope has always been a imaging scope and I used a meade zero shift focuser. No need to touch the focuser knob once set.

I removed the screws and guess what, the focuser is now nice and smooth.
I can now also see a noticeable movement in the mirror when the focuser is turned fully in and out.

I can now get to focus. This has solved my original problem.

However, I now have a collimation problem. I am going to start a new thread for this.

Thanks to everyone who offered help.

Brendan