No worries, Brent. The most sensitive component is the main board - use gloves.

A few more pics. Reconfiguring the cooling curve. It's a little different to the 1000D and needs more energy to achieve setpoint.

Decoding the OLED output most of which is for troubleshooting and of no value otherwise.

TMP - the analog value of the cold finger, which should match the TGT. Higher resolution than Celcius.
AIR -Celcius
SET - Celcius
CF - cold finger Celcius
DIFF - analog difference TMP and TGT
LO - TEC power off milliseconds
HI - TEC power on milliseconds
>T< on setpoint, otherwise UP> or <DN handy for aging eyes

A bit of cosmetic work to be done and a clean.